PM2010
|
|||||||||||||||||||||||||||||||||||||||||||||||||||||||||||||||||||||||||||||||||||||||||||||
| The PM2010 is a four-quadrant
power source capable of supplying up to 20 V/10
Amps to a load. The maximum power that may be
safely generated is 50 Watts on a continuous
basis and 200 Watts peak on a pulsed mode for no
more than 100 ms, and a duty cycle of less than
10%. The PM2010 uses an onboard controller, which relieves the system CPU of performing many functions necessary for the module operation. The result is higher efficiency of the CPU and faster execution of the module commands. All measurements use the Zero Time Average (ZTA) circuit (patent pending) that allows for a measurement time of less then 250 us to the full 16-bit resolution. ZTA enables the PM2010 to obtain the repeatability of using 20 to 50 measurements and averaging the results, with a single measurement. The PM2010 implements a calibration method called Hardware Error Correction (HEC) (patent pending). This method performs a full calibration, correcting errors for every code, as opposed to the more traditional two-point calibration of offset and gain correction. This method uses no CPU time, which accelerates the forcing and measurement functions. |
| Force Mode Specifications* | Measure Mode Specifications* | ||
| Maximum Output Voltage | ±20 V | Maximum Voltage | ±20 V |
| Maximum Output Current | 10 Amps | Maximum Current | 10 Amps |
| Maximum Power Dissipation | 50 W Continuous/200 W Pulsed | Input Impedance | 1 Meg Minimum |
| Programming Resolution | 14 Bits | Measure Resolution | 16 Bits |
| Current Ranges | 100 nA | Current Ranges | 100 nA |
| . | 1 uA | . | 1 uA |
| . | 10 uA | . | 10 uA |
| . | 100 uA | . | 100 uA |
| . | 1 mA | . | 1 mA |
| . | 10 mA | . | 10 mA |
| Voltage Ranges | 1 V | Voltage Ranges | 1 V |
| . | 2 V | . | 2 V |
| . | 10 V | . | 10 V |
| . | 20 V | . | 20 V |
| Accuracy | Accuracy | ||
| Current | ±(0.1%
of Range + 50 nA) for 100 nA to 100 uA Ranges ±(0.1% of Range + 300 nA) for 1 mA to 100 mA Ranges ±(0.5% of Range + 30 uA) for 1 A to 10 A Ranges |
Current | ±(0.1%
of Range + 50 nA) for 100 nA to 100 uA Ranges ±(0.1% of Range + 300 nA) for 1 mA to 100 mA Ranges ±(0.5% of Range + 30 uA) for 1 A to 10 A Ranges |
| Voltage | ±(0.05% of Range + 500 uV) | Voltage | ±(0.05% of Range +500 uV) |
| Settling Time | <1 ms | Acquisition Time | 250 us |
| Range Change Time | 15 ms | Range Change Time | 15 ms |
| Programming Time | 20 us | Programming Time | 20 us |
| Connect Time | 15 ms | Connect Time | 15 ms |
| *All specifications are subject to change without notice. |